Coronary artery overexpansion increases neointimal hyperplasia after stent placement in a porcine model
Robert J Russo1, Patricia D Silva1, Mark Yeager1,2
1 Division of Cardiovascular Diseases, Scripps Clinic, La Jolla, CA, USA
2 Department of Cell Biology, The Scripps Research Institute, La Jolla, CA, USA
ABSTRACT
Objective: To determine if an aggressive approach to coronary revascularisation with oversized balloons is counterproductive, we studied the effect of increasing balloon-to-artery (B:A) ratio on neointimal hyperplasia following primary stent placement using a non-atherosclerotic porcine coronary overstretch model.
Methods: 60 vessels in 33 Yorkshire swine were randomly assigned to one of five B:A ratios between 1.0:1 and 1.4:1. Intravascular ultrasound (IVUS) imaging was performed before bare-metal stent placement to accurately determine vessel size, after stent placement, and at 28 days.
Results: The mean prestent vessel diameter was 3.05 (0.31) (SD) mm. In-stent neointimal volume, in-stent volume stenosis and cross-sectional area stenosis at the stent minimum lumen diameter increased significantly with increasing achieved B:A ratio (multilevel regression test for slope, p<0.001, p = 0.002 and p<0.001, respectively) and were independent of vessel size. Even minor vessel overstretch at an achieved B:A ratio of 1.1:1 resulted in significant neointimal hyperplasia. Larger B:A ratios were also associated with more neointima beyond the stent edges (p = 0.008). For vessels from the same animal, neointimal response at a given B:A ratio was dependent upon the animal treated.
Conclusions: In a porcine model of IVUS-guided coronary primary stent placement, vessel overexpansion is counterproductive. Neointimal hyperplasia at 28 days is strongly associated with increasing B:A ratio. In addition, vessels do not respond independently of each other when multiple stents are placed within the same animal using a range of B:A ratios.
Heart 2007; 93: 1609-1615. |
|